Hi Dietrich,
Here is a histogram of the maximum pixel values of ESO_CCD_#66, the one the
was replaced that had two hot columns. All maximum values are above 60000
ADU and 25 of them are at the maximum possible of the data: 65535 ADU.
These are taken from 1345 ILT images of all types. I can look at all the
other CCDs to see if they are similar. Hopefully sometime this
afternoon we
will have a large amount of the commissioning data thus far ingested into
our system and be able to do a similar investigation on that data from here.
Cheers,
-=John
On Mon, 28 Mar 2011, Dietrich Baade wrote:
> Hi Gijs, John,
>
> we have run into some apparent problem with the OmegaCAM science mosaic, and
> Konrad amd Ewout suggested that I contact you.
>
> What we see
> Bright stars saturate well below 2^16-1 ADU. Often, this happens at 57,000
> ADUs; only the brightest stars reach 62,500 ADU. The cores of such images
> have a flat plateau. This is true for all CCDs (quite a few) in the mosaic
> that I've checked.
>
> What else we know
> The CCD test reports from e2v show full-well capacities of between 220,000
> and 340,000 e-/pixel. You have previously determined gains of 2.2-2.4
> e-/ADU. A quick on-line estimate here suggests ~2, which I consider
> consistent. A gain of 2.3 e- would imply 150,000 e- @ 65,535 ADUs, well
> below physical full well. FORS2 CCDs from the same batch have a gain of 2.4
> and don't run into physical saturation before the limit of the A/D converter
> is reached.
>
> What we suspect
> The numbers for OmegaCAM do not match. Since the entire science mosaic
> seems affected, the problem must be basic. It could be a wrong gain setting
> by S/W or H/W or similar. Olaf had no time yet to look into this.
>
> What could help us
> It would be great if you could search your archive for images containing
> saturated pixels with a value of 65535 (=2^16-1). Don't send us such
> images. It will be sufficient if you could provide a brief description.
> Ideal would it be if you found a CCD with a (partial) hot column. In that
> case, give as us the location or name of the chip and the column number.
> The faulty CCD that was replaced in 2009 might also have shown such
> saturated columns.
>
> If you do not find anything quickly, suspend your work and let us know how
> you've proceeded. We would, then, tell you whether it makes sense to
> continue.
>
> Many thanks in advance for your help.
>
> Best regards,
>
> Dietrich.
>
> --
> Dietrich Baade
> ESO - European Organisation for Astronomical Research in the Southern Hemisp
> here
> Karl-Schwarzschild-Strasse 2, 85748 Garching bei Muenchen, Germany
> Email: dbaade at eso.org Tel: +49 89 3200-6388 Fax: +49 89 3202362
>
>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ccd66_imstat_max.png
Type: image/png
Size: 15134 bytes
Desc: not available
Url : http://listman.astro-wise.org/pipermail/comm2011/attachments/20110329/5745a06d/attachment-0004.png